Hi, Please i need your supports, and get clarified the below ,
Refering Below extract from API 6A
A minimum of nine thermocouple test locations shall be used for all furnaces having a working zone greater than
0,3 m3 (10 ft3).
For each 3,5 m3 (125 ft3) of furnace working zone surveyed, at least one thermocouple test location shall be used,
up to a maximum of 40 thermocouples.
My intrepretation for TUS of furnace with volume 53 cubic meters working zone was 16 thermocouples(ie, 53/3.5=15.14 No.s) to be used during calibration, an auditor explained the same in another way that 16 + 9 a total of 25 No.s Thermocuples shalll be used.
Please i need your supports, and get clarified.