Hi, Please i need your supports, and get clarified the below ,

Refering Below extract from API 6A

A minimum of nine thermocouple test locations shall be used for all furnaces having a working zone greater than
0,3 m3 (10 ft3).
For each 3,5 m3 (125 ft3) of furnace working zone surveyed, at least one thermocouple test location shall be used,
up to a maximum of 40 thermocouples.


My intrepretation for TUS of furnace with volume 53 cubic meters working zone was 16 thermocouples(ie, 53/3.5=15.14 No.s) to be used during calibration, an auditor explained the same in another way that 16 + 9 a total of 25 No.s Thermocuples shalll be used.

Please i need your supports, and get clarified.

 
Replies continue below

Recommended for you

The number of thermocouples required is not additive. You are correct, 16 thermocouples are required.

Valvit
 
I don't know how closely API tracks AMS 2750, the pyrometry standard for aircraft heat treating.

Industrial Heating ran this table in 2010 for the D version of 2750 (current version is E) for number of thermocouples based on a volumetric calculation.

The table splits the auditor's quantity of 25 and your quantity of 16 at a middling 20 (@57m3)

thermocouple_locations_for_TUS.jpg


The original article and table are still on-line at
Link

I'm told that in the aircraft sector the auditor wins these kinds of arguments.
 
thanks alot.. now can we use production furnace thermocouples for TUS ??
 
Please tell me this wasn't an API auditor. I have seen some stupid stuff from auditors over the years, but this one is really over the edge.

I don't see how the auditor could win this one.

Consider a sign at a gas station:
Minimum charge for any amount over 1 gallon: $9.00
Charge for gasoline shall be $3.00 per gallon.

So, according to this auditor, if you bought 6 gallons, you'd have to pay $27 (instead of $18)?

now can we use production furnace thermocouples for TUS ??
Are you asking if you can use spare production T/C's for your survey?
I have never seen it done, but there is no reason it couldn't be done (provided proper documentation).

In general, however, the survey T/C's are Type K (Chromel/Alumel) and the production T/C's are Type S (Platinum/Pt-Rh). The cost of one Type S T/C could pay for a whole whole spool of Type K. It is common to run the survey with a few extra T/C's just in case one of two of them should fail during the survey.
 