Continue to Site

Eng-Tips is the largest engineering community on the Internet

Intelligent Work Forums for Engineering Professionals

  • Congratulations MintJulep on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Heating of the RTD wire leads 3

Status
Not open for further replies.

JoseDRG

Mechanical
May 21, 2010
12
Hi,

I'm working on a temporary fixture that will use an RTD to measure the temperature of a flat surface. The design might allow for the wires to the RTD to heat up to about 300F. Will this effect the performance/accuracy of the RTD significantly?

Any insight will be helpful.
 
Replies continue below

Recommended for you

Depends on the relative resistances, don't it?

TTFN

FAQ731-376
 
I doubt it.
First, the allowable temperature of the wiring to the RTD probably limits the temperature of the RTD environment but heat transfer from the wires should be miniscule.
 
50 ft distance using 30 gauge copper adds 10.3 ohms

230°F rise changes that resistance by 5.3 ohms

5 ohm error in 100 ohm platinum RTD is about 2°C error

TTFN

FAQ731-376
 
Thanks for your help however, I do have another questions.

The RTD is not fixed to the surface that it is measuring; contact will be done with slight pressure and no thermal conductive epoxy between the RTD and measured surface will be used. Can I expect accurate results?

Thanks for your help.
 
Thermocouple T (range -328 °F to 752 °F) could be an option. There are available models specifically conceived for flat surfaces.
 
most modern rtd's are flat unless they've neen encapsulated

how to you intend to account for radiation effects in your measurment?
 