KenAlmon
Electrical
- Apr 12, 2002
- 48
Where an instrument is supplied by a 24 Vdc, 2 wire cable, and information from the device is sent back over those two wires (as in a 4-20 mA loop for a temperature transmitter, or perhaps a digital "Profibus" signal), is there a code requirement for an equipment grounding conductor? If so, could you point me to the right section in the NEC? Is it voltage driven (24 Vdc OK but 48 Vdc and above not) etc.
My understanding from a power perspective, is that if the device could potentially "short out", then the case could become live (albeit at "only" 24-28 Vdc). If the case was plastic, then you could argue a ground is not required. However, if a junction box is used to make the terminations in, and the box was metal, you could probably argue that the box should be bonded back to the supply.
Complicating the issue is the shield, which, depending on practices, may or may not be present. However, the standard practice when shields are available are to float them at the field connection, which does not allow for bonding the box or instrument back to the supply.
Any NEC/IEEE/IEE/IAS comments appreciated.
Thanks
Ken Almon
My understanding from a power perspective, is that if the device could potentially "short out", then the case could become live (albeit at "only" 24-28 Vdc). If the case was plastic, then you could argue a ground is not required. However, if a junction box is used to make the terminations in, and the box was metal, you could probably argue that the box should be bonded back to the supply.
Complicating the issue is the shield, which, depending on practices, may or may not be present. However, the standard practice when shields are available are to float them at the field connection, which does not allow for bonding the box or instrument back to the supply.
Any NEC/IEEE/IEE/IAS comments appreciated.
Thanks
Ken Almon