Probes and floating measurements
I have been trying to find a good solution to the problem most of my life. Batteries? No, not really. But if needed, they have to last l-on-g. CMRR? Yes. Need that. High voltage and cat III? Yes, at least 600 V RMS and 100+ A fuses. Sometimes, the Isc is around 10 to 50 kA. Bandwidth? Seldom more than 10 MHz needed.
What do I use?
1. A very old Fluke DP120. With an extra battery pack (6x1.5 V AA) glued to it. 1200 V peak and 20 MHz. Runs for days between battery change. Problem: I only have one and I don’t trust Fuke any more.
2. Scope with more channels and measuring with ground as a reference. Careful probe compensation and A-B on screen. BW is then as good as your probe+scope. If scope is several times probe BW, you can forget the scope. Resulting BW is SQRT(sum of times squared).
3. Battery scope. That seems to be the ideal solution. But CMRR is terrible when you get up to MHz. Can’t be used. Tried several Fluke handheld scopes. Don’t work well when observing VFD outputs. Not even the C&A OX7204 can handle that kind of measurements without problems. OK, better than the yellow ones. But still.
4. USB scope with built-in differential probes. Best solution for the limited budget. I use a PicoScope 4444 with three differential inputs (8x4 mm banana sockets) and one “ordinary” BNC channel. Diff probes go to 10 MHz and BNC to 20. Suits me fine.
5. For high Isc, I use fused (and current limiting) strips that can handle 50 kA at 500 V RMS. That is OK for me.
Now looking at little A/D:s with Bluetooth. Batteries? Yes. But low drain and no CM problems whatsoever. Will probably use those when they start getting as good as I hope for. Maybe they are there already? Any input?