I have a board I designed that I'd like to improve but I'm fresh out of ideas. I'm wondering if anyone has any suggestions for what I think is an in-frequent but probably not uncommon problem.
I have a customer probe in a chamber that monitors the plasma within.
Typically the signal is brought in and buffered via some relatively quick amplifier. The amplifiers have obscene input resistances like 1016ohms so the probe is completely unloaded.
This is good because the probes have horrible source impedances that vary all over the place but tend to be between 700k and 1.5M ohms. Nasty!
If this was all there was to it I'd be thrilled. I'd be getting the required unsoiled reading owning to the extremely high amp input impedences.
The problem...
Frequently the plasma is turned off. This results in infinite impedance! So now the amplifier has a floated input. It promptly drifts up to the rail as does its output.. The solution is to put in a resistor from the input to ground.
Bye bye 1016.
A 1M pull down resistor up against a 1M sensor probe results in a loss of 50% of the signal. They need to know the DC levels here so I can't capacitively couple to the probe.
Looking for ideas on how to prevent leakage current from railing the amp while minimizing any loading on the sensor.
Keith Cress
kcress -
I have a customer probe in a chamber that monitors the plasma within.
Typically the signal is brought in and buffered via some relatively quick amplifier. The amplifiers have obscene input resistances like 1016ohms so the probe is completely unloaded.
This is good because the probes have horrible source impedances that vary all over the place but tend to be between 700k and 1.5M ohms. Nasty!
If this was all there was to it I'd be thrilled. I'd be getting the required unsoiled reading owning to the extremely high amp input impedences.
The problem...
Frequently the plasma is turned off. This results in infinite impedance! So now the amplifier has a floated input. It promptly drifts up to the rail as does its output.. The solution is to put in a resistor from the input to ground.
Bye bye 1016.
A 1M pull down resistor up against a 1M sensor probe results in a loss of 50% of the signal. They need to know the DC levels here so I can't capacitively couple to the probe.
Looking for ideas on how to prevent leakage current from railing the amp while minimizing any loading on the sensor.
Keith Cress