6
I am looking for instrumentation to measure oxygen consumed by primates during exercise. We are unable to attach the device to the monkey so we will construct a containment vessel.
 
Replies continue below

Recommended for you

Two O2 sensors. One on the air inlet and one on the air outlet. Solve for the delta. Set it up so you can, with the flip of a lever, swap which is sensing which outlet. Then drift and calibration can be offset easily.

Keith Cress
Flamin Systems, Inc.-
 
The inlet / outlet measurement also requires measuring the air flow rate. I suspect that the total error in the three signals and the associated calculations may be imprecise.
 