Is it Correct to calculate six sigma parameters in process that are controled by a PLC.
Example i use an Oven where i steriliza ampoules, i read with dataloggers the temperature inside oven´s chamber.
Must i calculate the Cp, Cpk and others measures of capability process?

 
Replies continue below

Recommended for you

Only if you want to keep it "in control".....

Remember, you must ensure that the reading are valid. A major problem I see is that people blindly believe anything with a digital readout. Just because it reports 156º, that doesn't mean it is!
 
Do a gage r and r before you start any great datalogging exercise.

Cheers

Greg Locock