Continue to Site

Eng-Tips is the largest engineering community on the Internet

Intelligent Work Forums for Engineering Professionals

  • Congratulations KootK on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

RS232 Cable Problems

Status
Not open for further replies.

lightl

Mechanical
Feb 8, 2016
42
For people who are using RS232 for Data transfer

Please stick to pin configuration when attaching the cable. 90% of the time, LSK error or DR signal off error comes due to improper soldering or wrong pin configuration.

25 pin on machine side and 9 pin on PC side.

I've attached the pin configuration for reference.

If data transfer is not occurring(LSK error)

1. Settings might be a problem: Check the software settings, Baud Rate Stop Bits, etc. It should match the machine parameters
2. Rs232 cable might be a problem: Always stick to pin configuration and check the cable for continuity.
3. Port problem on the machine side or PC side.: This happens very rarely but cannot be ruled out.

Also try using a USB to RS232 converter for better reliability.

 
 http://files.engineering.com/getfile.aspx?folder=f2534e92-31b2-46be-b97f-902260fe4acd&file=New_Bitmap_Image.jpg
Replies continue below

Recommended for you

IR, I initially had the same question... my guess is he was running into issues with the comms to his laser.

Dan - Owner
URL]
 
@IRStuff many old laser machines use rs232 comm ports.
 
the older bystronics used 2 com ports. One for machine/human interface and the other for moving job data

Chris Krug Maximum Up-time, Minimum BS
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor