Stent parallel plate compression

Status
Not open for further replies.

Bj3245

Bioengineer
May 16, 2024
3
I am modelling a parallel plate compression for a vascular stent in ABAQUS. I used 1/2 of the model as reported in literature, applied simmetry where the stent is cut, put a shell plate on the top with fixed displacement anb constrained JUST ONE NODE of the stent along the axial direction. However, I get this message: "The strain increment has exceeded fifty times---"
With one design everything went well, with this other one (same assembly, BCs, everything...) my job does not complete. Do you have an idea of why this happens? Here is the file (I tried to put 1mm of distance between the plate and the stent just to see if something changed).
Thanks in advance
 
Replies continue below

Recommended for you

You should model the plate as rigid - apply a rigid body constraint to it or create an analytical rigid part and apply boundary conditions to the reference point.

Also, don't use perfect plasticity for the stent. At least add some hardening.
 