I learned a valuable lesson the other day. I'm not electrical but work with many issues related to it.
We have a submersible motor we use and it requires
that we take ohm readings to checkt the integrity of
the moisture sensors. I'm suppose to get 330k ohm
with a +/- 10% variance. Within that range the
sensors are working within acceptable parameters.
But periodically I get in a rush and can't find my
clamps so I will hold the red and black leads in my
hands (acting like a clamp) against the two leads on
the sensors. Well I got a low reading of about 279k.
So I called the motor engineering center to ask about
this reading.
He said not to hold the leads with my fingers as this
would give me a low reading since it would be going
through my body. Do you guys agree? Not that I'm
second guessing this guy but I just thought I toss
this out to see if other's have experienced this as
well.
It may make a good Helpful Tip to put out. It's so
simple that anyone could make this mistake.
Wayne E. Lovison
service-parts@naglepumps.com
We have a submersible motor we use and it requires
that we take ohm readings to checkt the integrity of
the moisture sensors. I'm suppose to get 330k ohm
with a +/- 10% variance. Within that range the
sensors are working within acceptable parameters.
But periodically I get in a rush and can't find my
clamps so I will hold the red and black leads in my
hands (acting like a clamp) against the two leads on
the sensors. Well I got a low reading of about 279k.
So I called the motor engineering center to ask about
this reading.
He said not to hold the leads with my fingers as this
would give me a low reading since it would be going
through my body. Do you guys agree? Not that I'm
second guessing this guy but I just thought I toss
this out to see if other's have experienced this as
well.
It may make a good Helpful Tip to put out. It's so
simple that anyone could make this mistake.
Wayne E. Lovison