temp sensor testing

Status
Not open for further replies.
Replies continue below

Recommended for you

Since
- the D2401-2 trimode (triple mode) device is a level/interface and temperature measuring device
- one of the tools used is a tape/ruler with a traceability cert
- the units in that column are centimeters

then I would deduce that the measurement of an actual level at 93.0cm had an error of -0.3cm at a reported reading of 92.7cm.
 
Makes, much more sense now, BUT I would have trouble trusting an entity that uses "cms" for ANYTHING. After all, they didn't use "°Cs" in the other four columns.

Seems odd that they only tested one point for tape measure, maybe because the OP seems to think it's only a temp sensor

TTFN (ta ta for now)