Temperature Correction Formula

Status
Not open for further replies.

nde3rich

Petroleum
Jul 29, 2008
23
Can anyone tell me the formula used to manually correct ultrasonic thickness measurements for elevated temperatures ?
 
Replies continue below

Recommended for you

nde3rich,

According to Krautkramer - Ultrasonic Testing of Materials pg 14, the longitudinal velocity (which is applicable to thickness gages) decreases with a temperature increase by about 1 m/s per degree C on average up to 1200 C. Using the nominal longitudinal velocity for steel of 5,900 m/s; the thickness reading would be in error (thicker than actual) by .0169% for each degree C the test piece is above the calibration standard.

JR97
 
thanks, but I was looking for the expanded version.
 
ASME SE797 states a rule-of-thumb... for every 100F increase in temperatrure, the plate thickness measures approximately 1% thicker.

Joe Tank
 
All the above is essentially correct unless the temperature of your transducer is changing. There has to some means keep the transducer temperature near the calibration temperature.
 