So the emission curve follows a known form, fourth order I believe.
They don't emit anything.
They just measure in two bands.
The software knows that the ratio of these should exactly tell you the temperature.
By using bands that are fairly close you can assume that e is the same for both and that any blockage (occlusion, smoke, fumes) will have equal impact on both.
We used these for melting high temp alloys since the surface of melt was always changing and fumes were an issue.
Much better than old total radiation units.
There is also a very clever unit that uses a sapphire rod with the end cut at 45deg and gold plated.
They then connect it to a fiber optic bundle and do the same ratio math.
The beveled end gives you total internal refection and it follows the Black Body laws almost perfectly.
You could look at type C or D thermocouples, don't use type G.
These require a very low oxygen and low H2O environment.
I have only ever seen then used in pure hydrogen or high vacuum (<0.1 micron).
= = = = = = = = = = = = = = = = = = = =