Wild (Leica) distomat error 55

Status
Not open for further replies.

bobr57

Mechanical
Sep 5, 2009
2
I have a Wild DI1600 attached to a Wild T2002. When I request a DIST measurement it takes a while but eventually displays an error 55 on the Distomat (error 255 on the T2002).

I don't have a manual for the Distomat and the manual for the T2002 does not list this error code.

Can anybody tell me what I need to do to clear this?

Can anybody point me at an online copy of the DI1600 (or any other Wild Distomat) manual?
 
Replies continue below

Recommended for you

Try this.


55 appears to be - Interference during measurement (e.g no return signal, break in beam, excessive air turbulence)
Close range signal too weak.

press 1 ce 1; and or increase no of prisms; check adjustment of head

Measurement exceeds ambiguity. press 1 ce 1: set LDIL mode

Cheers,
Chris
 
Further to above , check out page 3 of the link especially wrt adjusting the DISTOMAT head

Cheers,
Chris
 
Chris,
Thanks for the tips.

I set up again with the prism about 100 ft away and still could not get a distance. Test mode indicated no return signal.

I moved the prism to about 20 ft -- still no luck.

I took the top cover off the distomat and looked around; nothing appeared to be amiss (pristine clean inside). The one (apparently) moving part seemed to move freely (looks like it is a shutter). I put the distomat back on the theodolite with the cover loose and observed the shutter (??) and low and behold a got a good distance result. I checked the distance with a hand held disto and got agreement within a few mm. YEA!! I took the distance several times and then took the distomat off to put the screws back in the cover.

Upon reattaching to the theodolite, I'm right back where I started: error 55. Tried taking the cover off again but no luck. Tried removing and reattaching the distomat several times to see if it's a bad connection to the theodolite; still no luck.

While it is nominally taking a distance, I can hear a very high pitched whine; is this normal? Any other ideas?

Thanks,
Bob
 
From memory I think that the shutter may be a "cascade filter" and it is probaly this which is malfunctioning.
I don't think that you can avoid a Dealer Service.%-(

Cheers,
Chris
 
Hi, i'm getting the same error on a TCRA1103 total station. besides when i switch the laser pointer on the laser won't ocme out of the telescope, however when i look through the front lens of the instrument i can see the laser cominfg out of the bottom of the telescope. do you think mine has the same thing?
 