lurker
Mechanical
- Aug 18, 2004
- 9
Hi out there,
I have been working on an issue for a while now and I think I have reached my wits end here... I hope someone out there can at least point me in the right direction. Let me see if I can explain this:
I have an issue which involves a cable that consists of two legs joined together zip-cord style (a little webbing in between holds them together). The cross section would look like " OO " essentially. Outer jacket is Santoprene TPR (rubber).
IEC 60601-2-4, Section 10, clause 56.101, subclause b. applies here.
We have a requirement on this product that the cable must withstand 10,000 cycles of flex, through 90 degrees each side of vertical (180 degrees total), 30 cycles per minute, a 5 N weight is suspended on the loose end of the cable, 300 mm from the axis of rotation. After 5000 cycles, the cable is turned 90 degrees and flexed another 5000 cycles in this new orientation (total is 10,000 cycles).
The cable will pass the "easy" bend, meaning a bend along the wide direction of the double cable, but when turned 90 degrees and the bend is in the against the wide direction of the cable, we are getting failures of the high voltage wires inside. The cable tends to fold over in this "hard" direction.
Any ideas on what can be done to help the strain relief/cable pass? Any proven strain relief geometries/designs out there for this double-cable configuration? Any ideas on whether changing the materials or cable configuration would help?
Right now the high voltage conductor (which fails) is made of Tinned Cadmium Bronze/Copper (C16200, "soft" or annealed) and is #22 AWG (26/36) construction. We need this cable to carry signal as well as deliver energy (defibrillation). Also, this product can be subject to steam autoclave at 270 deg F... While the jacket and insulation may protect it, our experience is that over time there is some penetration of the steam to the conductor.
What can we do to help this cable pass the IEC flex test?
Thanks in advance for any help that can be provided.
- lurker
I have been working on an issue for a while now and I think I have reached my wits end here... I hope someone out there can at least point me in the right direction. Let me see if I can explain this:
I have an issue which involves a cable that consists of two legs joined together zip-cord style (a little webbing in between holds them together). The cross section would look like " OO " essentially. Outer jacket is Santoprene TPR (rubber).
IEC 60601-2-4, Section 10, clause 56.101, subclause b. applies here.
We have a requirement on this product that the cable must withstand 10,000 cycles of flex, through 90 degrees each side of vertical (180 degrees total), 30 cycles per minute, a 5 N weight is suspended on the loose end of the cable, 300 mm from the axis of rotation. After 5000 cycles, the cable is turned 90 degrees and flexed another 5000 cycles in this new orientation (total is 10,000 cycles).
The cable will pass the "easy" bend, meaning a bend along the wide direction of the double cable, but when turned 90 degrees and the bend is in the against the wide direction of the cable, we are getting failures of the high voltage wires inside. The cable tends to fold over in this "hard" direction.
Any ideas on what can be done to help the strain relief/cable pass? Any proven strain relief geometries/designs out there for this double-cable configuration? Any ideas on whether changing the materials or cable configuration would help?
Right now the high voltage conductor (which fails) is made of Tinned Cadmium Bronze/Copper (C16200, "soft" or annealed) and is #22 AWG (26/36) construction. We need this cable to carry signal as well as deliver energy (defibrillation). Also, this product can be subject to steam autoclave at 270 deg F... While the jacket and insulation may protect it, our experience is that over time there is some penetration of the steam to the conductor.
What can we do to help this cable pass the IEC flex test?
Thanks in advance for any help that can be provided.
- lurker