I am trying to measure crosstalk between channels in a slip ring. Due to the assembly of the slip ring, there are 2-3m cables on each side of it, and my test signals have to go through these cables.
I am using a HP 8714B Network Analyser, in transmission mode, to measure the crosstalk.
I have connected the input of one of the slip ring channels to the RFout and have terminated the output of this slip ring channel (on the other side of the slip ring) with a 50 ohm load. I have then connected another slip ring channel to the RFin of the analyser, and have also terminated this channel with a 50 ohm load.
The input and output of the analyser are on the same side of the slip ring - the terminating loads are on the other.
Ok.. Here are my questions:
1) What are the effects of background noise on the crosstalk measurements. The tests are being performed on the factory floor, and there are several machines in operation. It is likely that the ground connection in the test area has various harmonics on it from this machinery, as well as there just being EMI in the area.. Would this be a problem, or would it couple to both channels of the analyser and hence not make a difference?
2) Related to 1, if I ground the case of the slip ring, might noise on the ground connection affect the crosstalk measurement?
3) The cables that connect to each side of the slip ring include several coaxial cables, and a few unshielded. I am testing between shielded channels... How much of an effect are the cables going to have on my crosstalk compared to the slip ring itself?
4) I think the worst of the crosstalk is being casued by a ground loop that connects the shields of the two channels I am testing... Is there anything I can do about this?
Sorry for such a long post and thank you very much for taking the time to read it. I would greatly appreciate any help and advice that you have to offer.
Best regards,
BPELEC.
I am using a HP 8714B Network Analyser, in transmission mode, to measure the crosstalk.
I have connected the input of one of the slip ring channels to the RFout and have terminated the output of this slip ring channel (on the other side of the slip ring) with a 50 ohm load. I have then connected another slip ring channel to the RFin of the analyser, and have also terminated this channel with a 50 ohm load.
The input and output of the analyser are on the same side of the slip ring - the terminating loads are on the other.
Ok.. Here are my questions:
1) What are the effects of background noise on the crosstalk measurements. The tests are being performed on the factory floor, and there are several machines in operation. It is likely that the ground connection in the test area has various harmonics on it from this machinery, as well as there just being EMI in the area.. Would this be a problem, or would it couple to both channels of the analyser and hence not make a difference?
2) Related to 1, if I ground the case of the slip ring, might noise on the ground connection affect the crosstalk measurement?
3) The cables that connect to each side of the slip ring include several coaxial cables, and a few unshielded. I am testing between shielded channels... How much of an effect are the cables going to have on my crosstalk compared to the slip ring itself?
4) I think the worst of the crosstalk is being casued by a ground loop that connects the shields of the two channels I am testing... Is there anything I can do about this?
Sorry for such a long post and thank you very much for taking the time to read it. I would greatly appreciate any help and advice that you have to offer.
Best regards,
BPELEC.