Validation and calibration of Welding equipment

Status
Not open for further replies.

kunjhon

Mechanical
Jun 15, 2020
12
Dear Expert,


Kindly define the difference between validation and calibration of the Welding machine.
 
Replies continue below

Recommended for you

Calibration involves having devices (volt meters and current shunt/meters) tested on the bench against know reference standards.
Validation is usually a simple function check that is done more often, but lacks the certificates.


= = = = = = = = = = = = = = = = = = = =
P.E. Metallurgy, consulting work welcomed
 
3.5 - verification
operations for the purpose of demonstrating that an item of welding equipment or a welding
system conforms to the operating parameters for that welding equipment or system

3.6 - calibration
set of operations which establishes, by reference to standards, the relationship which exists,
under specified conditions, between a DISPLAYED VALUE and a REFERENCE VALUE
[SOURCE: IEC 60050-311:2001, 311-01-09, modified – Notes removed and terms
"DISPLAYED VALUE" and "REFERENCE VALUE" used.]

3.7 - validation
operations for the purpose of demonstrating that a SET VALUE meets the REFERENCE VALUE
(within specified limits)

3.8 - consistency test
test to determine the repeatability of the equipment output over a period of time
 
ISO 17662

Calibration
Set of operations that establish, under specified conditions, the relationship between values of quantities indicated by a measuring instrument or measuring system, or values represented by a material measure or a reference material, and the corresponding values realised by standards

Validation
Confirmation through the provision of objective evidence that the requirements for a specific intended use (e.g. client specification) or application (e.g. product standard) have been fulfilled

Steve Jones
Corrosion Management Consultant


All answers are personal opinions only and are in no way connected with any employer.
 
Thanks for that Steve.

You saved me the trouble of saying something similar but less precise and eloquent.

In my (way back) shop experience, auditors were unduly concerned about welding power source calibration. Voltage meters never go out of calibration and ammeters never stay in it (and are not very useful anyway).

"If you don't have time to do the job right the first time, when are you going to find time to repair it?"
 