Coated electrode and pH reading

Status
Not open for further replies.

lapsd

Electrical
Mar 15, 2004
13
Hi alls,

What does it happen with pH reading when the electrode is starting to coat?

Thanks.
 
Replies continue below

Recommended for you

With a coated electrode the sensitivity of the electrode is diminished. In colloid suspensions you can expect erroneous readings.
m777182
 
One problem is the tiny ceramic pickup is prevented from absorbing the sample. Temperature is affected since the probe is now slightly insulated.

You can clean the probe with any industrial strenght cleaner. Just be sure not to scratch the glass.



Good Luck!