Continue to Site

Eng-Tips is the largest engineering community on the Internet

Intelligent Work Forums for Engineering Professionals

  • Congratulations KootK on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Differential Scanning Calorimetry (DSC)

Status
Not open for further replies.

jeanjmds

Materials
Oct 11, 2016
6
Hi,

I'm trying to use DSC to study phase transformation in a steel. I'm using an STA 449 F3 Jupiter equipment from Netzsch. The problem is that my equipment doesn't have vacuum and I saw some oxidation at the sample at the end of test. I'm using high purity argon (99.999%) as purge at 50 ml/min and I'd already tried to purge the chamber using argon for 1 hour before start test. My sample is 6 mm diameter and 0.2 mm tickness (between 10 and 20 mg as somebody told me and I saw in some books).

My question is: Does anybody knows how can I solve this oxidation problem in my sample without using vacuum?

Thanks.
 
Replies continue below

Recommended for you

1. Always keep you chamber under a light Ar purge, never just open to air.
2. Store samples in a container with descant to assure that they are very dry.
3. cycle the chamber empty before a run if you haven't used it for a while.
4. Install a dew point meter on the vent line, purge until you reach -40.

This is all in the details.

= = = = = = = = = = = = = = = = = = = =
P.E. Metallurgy, Plymouth Tube
 
That device is multi-purpose (TGA, DTA, DSC). If you use a tradition DSC, the sample is contained inside a sealed pan that will prevent oxidation.
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor