Hi,

A test performed on a flow meter indicate an uncertainty of 12% compared to the reference, knowing that the reference uncertainty is 5%.

Can we substract the uncertainty and assume that the real uncertainty of the flow meter is (12%-5%) 7 % ?

Thank you

 
Replies continue below

Recommended for you

No. Using that "logic", if your reference was 12%, you would have a really great accuracy from your meter.

You can only assume that your meter accuracy could be up to 12/5 = 1.4 x worse than your reference accuracy.

You meter will still measure +/-12% from the real flow.


--Einstein gave the same test to students every year. When asked why he would do something like that, "Because the answers had changed."
 
Thanks for your feedback,

Is there any standard to back calculate a flow meter "real" uncertainity taking into consideration the uncertainity of the flow meter used as a reference?
 
Well both of those numbers are truly terrible IMHO.

that's more like an educated guess than a meter.

Also normally you take a wide range of flows from 10% of max to 100% in 10% steps.

Your reference meter needs to be sent back and calibrated against a much much better meter and that uncertainty reduced to less than 1% for it to be of any use as a reference meter.

Or one or the other meters is operating at a small fraction of it's rated flow (<10%)

Remember - More details = better answers
Also: If you get a response it's polite to respond to it.
 
If your 12% meter reading was always within a certain percentage of the reference meter reading, say 2%, it would be statistically possible to infer that the 12% meter was 5+2 =7%, but it would be equally as possible to conclude the 5% meter has a worse accuracy than 5%, having an accuracy of 10-14%, only proving the old adage, "A man with two watches never knows the time."

--Einstein gave the same test to students every year. When asked why he would do something like that, "Because the answers had changed."
 
How was the reference uncertainty determined?

Remember - More details = better answers
Also: If you get a response it's polite to respond to it.
 
I thought you were talking about a properly calibrated and certified "standard reference meter" as in prover looped, or something. Tank volume measurements are never going to make a reference standard. Notoriously poor accuracy, temperature variations everywhere, pressure variations everywhere. How did you even do that? But it does explain why your reference is not better than 5%.


--Einstein gave the same test to students every year. When asked why he would do something like that, "Because the answers had changed."
 
What type of meter are we taking about here?

What's a separator got to do with it?

Is this two phase flow?

Remember - More details = better answers
Also: If you get a response it's polite to respond to it.
 
triviaux1 said:
The reference measurement is a two phase test separator and liquid measurement was cross checked using a tank
Note that part of separator liquid volume is occupied by popping up bubbles so gas-liquid level measurement and liquid density calculation are not simple issues.
 