Foreign object analysis of CMOS sensor via SEM/EDS

Status
Not open for further replies.
Replies continue below

Recommended for you

The analyses you presented suggests this is composed of nitrogen, sulfur, and calcium. It may well be organic but carbon is the overwhelming peak for exemplar and contaminant so you can't tell that way. Two things I would do is recollect the spectra down at about 8 keV to crank down the beam penetration, helping with contaminant identification. Semiquants are useful for this purpose of contaminant identification, but I would not include carbon at all - procedure notes carbon cannot be quantified due to its low atomic weight, so the numbers for it just confuses the issue.
 