Todd_A
Automotive
- Aug 30, 2016
- 5
I have read several posts that warn against doing GRR on flatness because of variation of the appraiser. How does that coincide with a robotic cell, where there is no human interaction? A robot loads the part on a CMM, and the flatness is scanned. The robot then removes the part, picks up another, etc. until the sample size is complete.
Run A: 0.0210 0.0242 0.0264 0.0224 0.0247 0.0281 0.0189 0.0221 0.0215 0.0236
Run B: 0.0206 0.0246 0.0258 0.0219 0.0246 0.0269 0.0191 0.0217 0.0212 0.0233
Run C: 0.0207 0.0241 0.0260 0.0217 0.0240 0.0273 0.0193 0.0214 0.0208 0.0226
These results are comprised of 10 pieces, each being ran three times. My total GR&R is coming in at 14.51%! Why? Clearly the variation between the ABC runs is minimal. My equipment variation is coming in at 11.27%. I don't understand why everything looks so poor when the actual variation is so tight. By the way, those numbers are metric.
Below is the advice being given by ANOVA. Furthermore, I tested the data in another program and got similar results.
"Your measuring equipment may need calibration or maintenance."
This is a brand new CMM. The GR&R was performed the day after installation was complete. I'm 100% positive, without question that calibration is good. I went as far as running a probe validation after each 10 piece set.
Can anyone offer some advice?
Thank you
Run A: 0.0210 0.0242 0.0264 0.0224 0.0247 0.0281 0.0189 0.0221 0.0215 0.0236
Run B: 0.0206 0.0246 0.0258 0.0219 0.0246 0.0269 0.0191 0.0217 0.0212 0.0233
Run C: 0.0207 0.0241 0.0260 0.0217 0.0240 0.0273 0.0193 0.0214 0.0208 0.0226
These results are comprised of 10 pieces, each being ran three times. My total GR&R is coming in at 14.51%! Why? Clearly the variation between the ABC runs is minimal. My equipment variation is coming in at 11.27%. I don't understand why everything looks so poor when the actual variation is so tight. By the way, those numbers are metric.
Below is the advice being given by ANOVA. Furthermore, I tested the data in another program and got similar results.
"Your measuring equipment may need calibration or maintenance."
This is a brand new CMM. The GR&R was performed the day after installation was complete. I'm 100% positive, without question that calibration is good. I went as far as running a probe validation after each 10 piece set.
Can anyone offer some advice?
Thank you