mrfailure
Materials
- Aug 7, 2011
- 809
In a recent thread we discussed differences in measured Vickers microhardness when using different loads despite the basis of the procedure as-published that there should be no difference. The April 2012 issue of Advanced Materials & Processes has a good paper by George Vander Voort and Ryan Fowler which addresses this very issue. I have attached the link to it. The upshot of the research is that perceived differences obtained at 100g loads and below are attributable to poor optics, and not differences in material or load. They found the perceived differences disappeared (at least for them) when using higher magnification objectives (i.e. higher than the 40X objective found on many indenters) and better lighting for measurement at low loads.
Aaron Tanzer
Aaron Tanzer