dear folks
Assalam O alaikum And GOOD NOON
PLZ INTERPRET THE CODE ASME BPV SEC V ARTICLE262
CZ I M CONFUSED DURING READING
FOR HOW MUCH PERIOD after DENSITOMETER SHOULD BE RE CALIBRATED FROM LAB?
T-262 Densitometer and Step Wedge
Comparison Film
T-262.1 Densitometers. Densitometers shall be calibrated
at least every 90 days during use as follows:
(a) A national standard step tablet or a step wedge calibration
(It mean after 90 Days)
T-262.3 Periodic Verification
(a) Densitometers. Periodic cablibration verification
checks shall be performed as described in T-262.1 at the
beginning of each shift, after 8 hr of continuous use, or
after change of apertures, whichever comes first.
(b) Step Wedge Comparison Films. Verification checks
shall be performed annually per T-262.2.
(now what ? after 8 hrs or 90 days or 1 year)
when will due the nxt calibration ?
regards
Jam Nauman
Assalam O alaikum And GOOD NOON
PLZ INTERPRET THE CODE ASME BPV SEC V ARTICLE262
CZ I M CONFUSED DURING READING
FOR HOW MUCH PERIOD after DENSITOMETER SHOULD BE RE CALIBRATED FROM LAB?
T-262 Densitometer and Step Wedge
Comparison Film
T-262.1 Densitometers. Densitometers shall be calibrated
at least every 90 days during use as follows:
(a) A national standard step tablet or a step wedge calibration
(It mean after 90 Days)
T-262.3 Periodic Verification
(a) Densitometers. Periodic cablibration verification
checks shall be performed as described in T-262.1 at the
beginning of each shift, after 8 hr of continuous use, or
after change of apertures, whichever comes first.
(b) Step Wedge Comparison Films. Verification checks
shall be performed annually per T-262.2.
(now what ? after 8 hrs or 90 days or 1 year)
when will due the nxt calibration ?
regards
Jam Nauman