LEB1
Materials
- Apr 23, 2012
- 15
Rockwell Hardness Testing to ASTM E18 :2014
Clause 5.2.4 states:
Rockwell hardness measurements shall be rounded in accordance with Practice E29 with a resolution no greater than the hardness values display of the testing machine.
we have recently upgraded our machine and the display resolution is to 1 decimal place. Technically we report to this level and have now been questioned from one customer who is requesting this rounded to the nearest whole integer. he is citing Note 2 of the standard which states that common practice for acceptance testing is to round to the nearest whole number, as we are a commercial testing facility we are not performing acceptance testing ( which I believe is related to production facilities )
my question is - if our Rockwell testing machine displays to 1 decimal place can we round to the nearest whole number and still be compliant with ASTM E18:14
thanks
Clause 5.2.4 states:
Rockwell hardness measurements shall be rounded in accordance with Practice E29 with a resolution no greater than the hardness values display of the testing machine.
we have recently upgraded our machine and the display resolution is to 1 decimal place. Technically we report to this level and have now been questioned from one customer who is requesting this rounded to the nearest whole integer. he is citing Note 2 of the standard which states that common practice for acceptance testing is to round to the nearest whole number, as we are a commercial testing facility we are not performing acceptance testing ( which I believe is related to production facilities )
my question is - if our Rockwell testing machine displays to 1 decimal place can we round to the nearest whole number and still be compliant with ASTM E18:14
thanks