I have question regarding our electrodynamic vibration shaker system.
Recently we have noticed some abnormal sound (like turning fork) from the system when it is used with head expander. It occurs at the frequencies,300 Hz, 600Hz, 900Hz, 1200 Hz, 1500Hz, 1800 Hz and so on. This sound appears even without the head expander and the situation get worsen with head expander.
We conducted, total harmonic distortion test at vertical axis with bare head expander as mentioned below.
Control acceleration = 1g
Test Frequency(Hz)
5
10
20
30
500
900
1200
1500
CH1 THD +N %
N/A (5Hz)
N/A (10Hz)
8.49286 (20Hz)
7.44264 (30Hz)
8.98874 (500Hz)
53.0711 (900Hz)
26.9406 (1200Hz)
9.44025 (1500Hz)
(CH2 THD + N %)
N/A (5Hz)
N/A (10Hz)
7.02918 (20Hz)
6.28316 (30Hz)
7.08433 (500Hz)
45.2265 (900Hz)
20.5199 (1200Hz)
9.21418 (1500Hz)
Control acceleration = 5g
Test Frequency(Hz)
500
900
1200
1500
(CH1 THD+ N %)
5.1278 (500Hz)
17.4902 (900Hz)
6.4243 (1200Hz)
2.0185 (1500Hz)
(CH2 THD +N %)
6.01514 (500Hz)
12.2896 (900Hz)
6.2920 (1200Hz)
2.7729 (1500Hz)
We think that above calculated THD values are very high and a clear distortion is visible in the graph of the head expander (pls see attached graphs).
Shaker manufacturer advised that system is working fine after reviewing these results.
If you can review this data and provide any feedback would be highly appreciated.
Thank you
BR,
Prasanna
I have question regarding our electrodynamic vibration shaker system.
Recently we have noticed some abnormal sound (like turning fork) from the system when it is used with head expander. It occurs at the frequencies,300 Hz, 600Hz, 900Hz, 1200 Hz, 1500Hz, 1800 Hz and so on. This sound appears even without the head expander and the situation get worsen with head expander.
We conducted, total harmonic distortion test at vertical axis with bare head expander as mentioned below.
Control acceleration = 1g
Test Frequency(Hz)
5
10
20
30
500
900
1200
1500
CH1 THD +N %
N/A (5Hz)
N/A (10Hz)
8.49286 (20Hz)
7.44264 (30Hz)
8.98874 (500Hz)
53.0711 (900Hz)
26.9406 (1200Hz)
9.44025 (1500Hz)
(CH2 THD + N %)
N/A (5Hz)
N/A (10Hz)
7.02918 (20Hz)
6.28316 (30Hz)
7.08433 (500Hz)
45.2265 (900Hz)
20.5199 (1200Hz)
9.21418 (1500Hz)
Control acceleration = 5g
Test Frequency(Hz)
500
900
1200
1500
(CH1 THD+ N %)
5.1278 (500Hz)
17.4902 (900Hz)
6.4243 (1200Hz)
2.0185 (1500Hz)
(CH2 THD +N %)
6.01514 (500Hz)
12.2896 (900Hz)
6.2920 (1200Hz)
2.7729 (1500Hz)
We think that above calculated THD values are very high and a clear distortion is visible in the graph of the head expander (pls see attached graphs).
Shaker manufacturer advised that system is working fine after reviewing these results.
If you can review this data and provide any feedback would be highly appreciated.
Thank you
BR,
Prasanna