Continue to Site

Eng-Tips is the largest engineering community on the Internet

Intelligent Work Forums for Engineering Professionals

  • Congratulations IDS on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

voltage drift

Status
Not open for further replies.

E2005

Electrical
Jul 18, 2005
46
I am using a strain gauge signal conditioner(120 ohm gauge)with another 120 dummy resistor gauge and rest of the bridge internally balanced by the chip(Analog device 1B31). Ideally the output should be zero if no force is applied on the bridge(balanced bridge) but Every time I start the circuit the output starts at a different value(value offsetted by +/- 0.4 or 0.5V from zero). I have to use a pot to bring it to zero everytime. But after a few minutes sitting idle it starts drifting again. I am using low pass(RC) filters at input and output. Someone suggested its the capacitor causing the drift (as it charges)so I tried different electrolytic capacitor values but hasn’t helped(need cutoff around 30 Hz).I am trying to find what’s causing the drift.
 
Replies continue below

Recommended for you

The amplifier shouldn't drift that much if it isn't damaged. If changing the amplifier doesn't help, then check the bridge output voltage with a DMM (microvolt resolution probably needed) or simply wire in a set of resistors to mimic the bridge. That should tell you if it is the amplifier or bridge/wiring that is the culprit.

Also, thermal EMF does just that. Creeping in a few minutes after power-on. Make sure that your wiring doesn't introduce unwanted EMF.

Gunnar Englund
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor