I have a strain gauge connected to a metallic rod hinged at one end. The system detects the bending of the beam(when I press it with my hand, as seen by increase in output voltage) but even before touching the rod, as I bring my hand close to the rod the output starts increasing (offsetting from zero point) and decreases to normal when I move away from it. The change in output voltage due to this offset is greater than that caused by bending. It seems almost as if the grounding of the system changes. I tried insulating and shielding the circuit, wires and beam(with aluminum foils, tapes etc) but still the systems offsets as soon as my hand or body comes close to the rod. I am trying to understand the reason for it and if there’s a way to solve this problem.
 
Replies continue below

Recommended for you

Are you grounding yourself while running these trials? Is it the same ground as the strain gauge? Is the rod sensitive enough to thermal effects such that bringing your hand within close proximity induces measurable strain?

Regards,
 
PSE: I did not ground myself to the rod because that is the way it will be used in the field, the person pressing the rod will not be able to ground himself to the rod every time he presses it. However apart from the person pressing the rod, everything else(strain gauge circuit, power supply, rod ) has a common ground. Also I do not know how sensitive the rod is to thermal effect( but I believe its not that sensitive that bring hand close to it would change its temperature or measurable strain)
 
My background is not electrical but I am still suspicious of separating the operator from the same ground as the device. Any way to test performance with a common ground? If it works you might be able to design a ground location on the equipment for the operator to contact while running a test.

Regards,
 
I think you need to make sure the strain gauge instrumentation works correctly when hooked up under lab conditions. Also, make sure your connections are good. Is it possible that you might have an "open circuit"?
 