Extensometer data vs. tensile strain

Status
Not open for further replies.

salma424

Materials
Feb 24, 2011
1
I have been trying to use an axial extensometer on instron 5560. The data that I get from using the extensometer shows as "strain 1%" is 20 times more than the actual data that I expect ( the data from crossheads moving which is "tensile strain").
How can I change that?
The gauge length of extensometer is 1 inch. The gauge length of the crossheads is about 1.5-2 inches.
 
Replies continue below

Recommended for you

Did you calibrate the extensometer? It sounds like the signal for turning voltage in the extensometer into specimen displacement/strain is not calibrated properly.
 
What type of extensometer is it?
Is it really meant for these long extensions?

2" is a very short distance between the grips. In elastomers there is very non-uniform deformation near the grips.
Are these straight uniform cross section samples?
Try a longer sample, say 8" between grips to start and see if the two values agree closer.

= = = = = = = = = = = = = = = = = = = =
Plymouth Tube
 
Not my expertise but; it sounds like mixing crosshead movement with extensometer data. And "gage length" applies to test coupon , not crosshead.
 