We are using the Mic 10 Portable Hardness Tester, we have a Test Block that has a certificate stating hardness was verified as 236 HV5, can we use the Mic 10 with HV10 probe to calibrate from this Test block?.
 
Replies continue below

Recommended for you

There should be a separate test block for the HV10 probe.

If you don't have that, then go ahead with what you have proposed with a possible 5% error.

DHURJATI SEN
Kolkata, India


 
If you're planning on using this in a professional capacity whereby all instruments/equipment need to be calibrated then you should purchase the HV10 calibration block.
 