Some process instruments such as PT (pressure transmitters) demonstrate an accuracy of 0.1% or 0.2% (depends on the technology of course). I understand that to calculate the overall measurement error of the measurement system (measurement system: PT + impulse line + Valves + PLC input module + Pressure calibrator), we need to take into account all sources of measurement errors.
We need to set pass/fail criteria for the measurement error to help the technician during calibration/testing of a particular instrument to tell if the measurement errors he records are acceptable or not. (the technician uses a certified Druck pressure calibrator)
Now considering the number of instruments we got, it would be a huge and tedious task to calculate the overall measurement error for each instrument using the usual error formula.
I would like to know if there is a rule of thumb to define the acceptable measurement error by type of instrument.
An example:
- The PTs we got on site have 0.2% accuracy
- The Druck calibrators we got have 0.025% accuracy
- The PLC input module has an accuracy of 0.1%
- Add a safe margin to take into account the remaining source of errors (Impulse line, elevation, etc.)
In this case, I would say that I can use 1 or 2% error as a good and practical pass/fail criteria for the measurement error.
What do you think ? Is my reasoning correct ?
Djaz
We need to set pass/fail criteria for the measurement error to help the technician during calibration/testing of a particular instrument to tell if the measurement errors he records are acceptable or not. (the technician uses a certified Druck pressure calibrator)
Now considering the number of instruments we got, it would be a huge and tedious task to calculate the overall measurement error for each instrument using the usual error formula.
I would like to know if there is a rule of thumb to define the acceptable measurement error by type of instrument.
An example:
- The PTs we got on site have 0.2% accuracy
- The Druck calibrators we got have 0.025% accuracy
- The PLC input module has an accuracy of 0.1%
- Add a safe margin to take into account the remaining source of errors (Impulse line, elevation, etc.)
In this case, I would say that I can use 1 or 2% error as a good and practical pass/fail criteria for the measurement error.
What do you think ? Is my reasoning correct ?