We are trying to measure tensile/compressive strains in 0.020" thick steel, using strain gages. With material this thin, the surface strains are dominated by local bending effects (e.g., oil canning). As usual, I average the top and bottom strains to get the neutral axis strain. I setup some cantilever bending pilot studies (method below) to verify that the bending effects cancel.
But in these pilot studies, my measured tensile (T) strains are consistently 5% to 10% lower than the compressive (C) strains when using the .020" material. (Pilots on 0.125" or 0.187" thickness steel show T and C match <1%.) Compressive strains match theory <1%, so the tensile strains are the error. Usually, the %error grows with the applied strain: at 200 applied microstrain (ue), T<C by 2-3%; %error increases to 5-10% at 500ue. All gages show some creep (typical for the CA adhesive): zero-return shifts 5-10 microstrain (1-2%) after applying 500 microstrain. We would like to avoid the epoxies if possible bc of the pot life or heat-cure requirements.
Here are some ideas, mine and Vishay’s, on why the tensile %error only shows up with the thin material:
1) Curved beam theory: Nope -- we aren’t near that territory.
2) The gages locally reinforce the material, and the polymer backing is stiffer in T than C. Vishay’s CEA gages are .003" thick; the EA gages are .001" thick. But my tests show EA gages are only slightly better than CEA gages for the %error.
3) The CA slips more when it is in convex curvature than in concave curvature. ?
4) The CA is slipping much more when under load, and more in T than C, than the zero-return values would suggest (visco-_elastic_). But shear is shear, right? Why would T be different than C, and error only show up with thin material?
5) ?
Details on the method:
1) I make a 1" x 8" x .020" steel strip "beam" (we have tried both stress relieved and as-rolled steel).
2) Mount a uniaxial gage on one side (or both), in the middle of the beam, using standard Vishay techniques, gages, instruments (MBond 200). I have tried various ways of clamping and other CAs but none solve it.
3) Precondition the gage by bending back and forth +/- 500ue and then lower strains.
4) Zero the meter by carefully holding the beam vertically (produces <1ue tension).
5) Secure one end of the beam and record the initial strain due to beam weight. Turn beam over and record strain due to opposite loading.
6) Apply a cantilever bending load to the end of the beam and record strain. I angle the mounted end of the beam up slightly so that the gage is roughly horizontal as the beam curves over.
7) Turn the beam over, mount identically as possible, and apply identical end load. (Results are not sensitive to these variations).
8) Repeat 5 and 6 with greater and greater loads, to about 600ue. I order it as T-C, C-T, T-C, C-T... to cancel order effects.
Thanks for any leads you might have!
David Malicky
But in these pilot studies, my measured tensile (T) strains are consistently 5% to 10% lower than the compressive (C) strains when using the .020" material. (Pilots on 0.125" or 0.187" thickness steel show T and C match <1%.) Compressive strains match theory <1%, so the tensile strains are the error. Usually, the %error grows with the applied strain: at 200 applied microstrain (ue), T<C by 2-3%; %error increases to 5-10% at 500ue. All gages show some creep (typical for the CA adhesive): zero-return shifts 5-10 microstrain (1-2%) after applying 500 microstrain. We would like to avoid the epoxies if possible bc of the pot life or heat-cure requirements.
Here are some ideas, mine and Vishay’s, on why the tensile %error only shows up with the thin material:
1) Curved beam theory: Nope -- we aren’t near that territory.
2) The gages locally reinforce the material, and the polymer backing is stiffer in T than C. Vishay’s CEA gages are .003" thick; the EA gages are .001" thick. But my tests show EA gages are only slightly better than CEA gages for the %error.
3) The CA slips more when it is in convex curvature than in concave curvature. ?
4) The CA is slipping much more when under load, and more in T than C, than the zero-return values would suggest (visco-_elastic_). But shear is shear, right? Why would T be different than C, and error only show up with thin material?
5) ?
Details on the method:
1) I make a 1" x 8" x .020" steel strip "beam" (we have tried both stress relieved and as-rolled steel).
2) Mount a uniaxial gage on one side (or both), in the middle of the beam, using standard Vishay techniques, gages, instruments (MBond 200). I have tried various ways of clamping and other CAs but none solve it.
3) Precondition the gage by bending back and forth +/- 500ue and then lower strains.
4) Zero the meter by carefully holding the beam vertically (produces <1ue tension).
5) Secure one end of the beam and record the initial strain due to beam weight. Turn beam over and record strain due to opposite loading.
6) Apply a cantilever bending load to the end of the beam and record strain. I angle the mounted end of the beam up slightly so that the gage is roughly horizontal as the beam curves over.
7) Turn the beam over, mount identically as possible, and apply identical end load. (Results are not sensitive to these variations).
8) Repeat 5 and 6 with greater and greater loads, to about 600ue. I order it as T-C, C-T, T-C, C-T... to cancel order effects.
Thanks for any leads you might have!
David Malicky