Hello. We have a movement on the issue. After replacing all boards and cables / SSIF, LASERIF, DIOSSI / we already have a result. The machine does not turn on and off when asked, but now cannot create a vacuum. Up to 30 mBAR. Now I expect BISTRONIC to send me new seals and I will replace all of...
Good afternoon, Mr. Krug. Today I looked at the CNC cabinet, and did not understand what exactly is the input module. I photographed what I guess it is. If this is the module, then on the first point about SHUTTER, I have to look at output one if it is indicated. Therefore, I have to run a cable...
Hello again. Before you answered me, I tried pressing the locale button and the chassis started at minimum power. I tried to cut, but there was no beam. Where can I find the schematic numbers? Is it possible for me to contact you via WhatsApp?
Hello Mr. Krug. Seeing one of your posts, I changed one of the items in the MMC menu from MCC / SSC to MCC, but the result is the same. The HV error disappeared but the chassis continued to go...
The ssif cable.
Maybe I need to change something in the software settings of the machine? Set the machine to run in manual mode?
Thankshttps://files.engineering.com/getfile.aspx?folder=3db2a00f-c1f0-4511-bd72-6abb357fb8cf&file=WhatsApp_Image_2020-08-03_at_08.57.41.jpeg
The board canfigurationhttps://files.engineering.com/getfile.aspx?folder=65c2f3e2-ba0f-4458-98ec-844b64d5cb1a&file=WhatsApp_Image_2020-08-03_at_08.57.40_(4).jpeg
The HV desk screenhttps://files.engineering.com/getfile.aspx?folder=6e3fa140-a5ce-42de-9c86-8934bdb7478b&file=WhatsApp_Image_2020-08-03_at_08.57.40_(1).jpeg
The hv desk errorhttps://files.engineering.com/getfile.aspx?folder=a27821fc-9db6-4ca9-830d-3b52ba8ce320&file=WhatsApp_Image_2020-08-03_at_08.57.40_(3).jpeg
Hello Mr. Krug. Separating the sssif board from the main board and putting it in step mode, the machine makes a mistake, the chassis does not start and there is no beam in the resonator
Hi, looking at the diagram, do I have to completely remove the board from the socket or just disconnect the cable? Attached are pictures of HV DESK, with the configuration of the boards.
Thanks again for your...